I have recently changed jobs and inherited some methods which were carved in stone and carried down the mountain long ago.
We have a method for volatile acids (C2 -C5) which involves aqueous injection of a 0.2%HCl extract of the acids. Concentrations are around 100-400ppm.
The biggest problem is carryover - we run more blanks than samples.
NOW FOR SOME DIAGNOSTICS:
--->If I run the 400ppm standard, followed by blank injections of 0.2%HCl in water, it takes about 5 blanks to clear out the carryover.
--->If I run a temperature profile with no injection - no ghost peaks appear.
--->This seems to suggest that the carryover is not residing on the column, but within the injector plumbing.
--->We are making a 1ul aqueous injection at 150C and 6.3psi. If my HP Flow calculator program serves me correctly I believe we are exceeding the internal volume of the 4mm single gooseneck liner we are using. Therefore the excess aqueous, acidic crud has to go somewhere....probably someplace bad.
--->If I give the system a front end overhaul, new liner, new seal, chop off a foot of column, the system performs marginally better for about 10 injections.
SO:
The first obvious thing to do would be to reduce injection size and stop overloading the injector.
The next step would be to move away from acidified aquous injections..but that's another post.
Any thoughts?
I would like to get some second opinions. I want to make some improvements, and I am up against a certain amount of inertia...