I've been a chromatographer for a quarter-century. Some things don't change with time, do they? This high-pressure/lo pressure battle is as old as HPLC, and never seems to end!
I've sold HPLC's and I've bought HPLC's, so I guess I've seen both sides of this particular issue.
First - You always need to degas your mobile phase, irrespective of high-press or lo-press. High-press vendors claim 'there's no need to degas'. Truth is, if you don't degas, outgassing can and will occur in the worst possible place - your detector's flow cell. The flow cell, remember, is at atmospheric pressure.
'Back-pressure regulators will eliminate outgassing'. Back-pressure regulators do just that - they put a significant back-pressure on the flow cell, thus suppressing outgassing. Unfortunately, quartz flow cells can crack at around 350 psi. Makes more sense to degas your solvents in the first place.
So, you degas. High-press or lo-press, you degas.
'Lo-press systems cost less' - they don't.
'Lo-press systems have lower dead-volume.' True, but not really an issue if you're using a good dynamic mixer for gradient mixing.
In fact, gradient homogeneity with lo-press systems can be a problem. This is because of the proportioning valves that lo-press systems use.
Bottom-line: It doesn't matter which system you prefer. What matters is the results and separation you finally get. Give or take a few theoretical plates, you'll get the same results eventually, provided your separation chemistry is OK to start with.