These last days I have been trying to set up a LC-ESI(-)-MS method to determine different acidic pharmaceuticals(pKa from 3.2 to 4.4).
As reported in many articles I am trying to work in a negative mode so I tried different mobile phases and check which one gave me the best signal. The best one was a gradient with Acetonitrile-Ammonium acetate 50mM pH=5.47 adjusted with formic acid. My standards are prepared in 2%acetic acid (due to my pre-treatment step).
The column I use is a C18 (15cm, 2.1cm, 3um) and the flow 0.2 mL/min.
Last Friday, I developed a gradient where I could separate the 3 compounds. However on Monday when I tried to reproduced this, it did not work. 2 of the 3 compounds eluted with the mobile phase and the third one eluted much earlier (about 15 min before). I tried different things, like change the mobile phase to acidic one and finally I reduced the flow to 0.1 mL/min. That way I could see the 3 peaks (all confirmed by MS). This morning I have run again the same sample at 0.1 mL/min, twice. I could see the 3 peaks, however retention time was decresed a little. Then I changed back to 0.2 mL/min and I see the 3 analytes. I repeat the injection and then again the first 2 analytes elute with the mobile phase and the third one elutes as Monday.
A colleague said he has seen similar behaviour and the best is that I have an isocratic gradient.
I do not know how to explain this and how to solve it.
Hope someone can help me to figure out what it can be...as I have never seen this problem befer in HPLC.