Area response as a function of flow

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

4 posts Page 1 of 1
I'm working on a new RP-LC method and noticed I'm getting different area responses at different flow rates? I'm actually seeing a 25% reduction in area when going from 1.5 ml/min to 2 ml/min. Why?

I can give other parameters is required, but it doesn't seem necessary.
Area is a function of the analyte absorbance and its residence time in the flow cell.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
To elaborate, peak area equals peak width times absorbance. Absorbance does not change when increasing flow rate from 1.5 ml/min to 2 ml/min, however the peak width decrease by a factor of 25% (1/1.5 - 1/2)*100/(1/1.5) =25%.
Ok that was very helpful. Thank you.
4 posts Page 1 of 1

Who is online

In total there are 327 users online :: 1 registered, 0 hidden and 326 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Semrush [Bot] and 326 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry