Help for chiral separation of a 2,3-dihydroxy-butyl ester.

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

7 posts Page 1 of 1
Hi, everyone:
I have the enantiomers of Benzoic acid 2,3-dihydroxy-butyl ester to be separated by chiral columns. I have ChiraDex (Merck), ChiroBiotic T (Astec), and ES-Pepsin (Agilent) columns in my hands. I have tried many moblie phases, such as the mixtures of H2O/MeOH, H2O/acetonitrile or H2O/EtOH, and many additives, such as isopropanol, ethylene glycol and diethylene glycol. And I also tried many runs at different temperature and different PH value.
But I still can not separate the two enantiomers by those three chiral columns.
Could you please give me some advices?
Do I have to order another kind of chiral column for the enantiomers?

Thanks a lot!

Chiral separations differ from other analyses in that most of the selectivity resides in the stationary phase, with the mobile phase having only a minor effect (the antibiotic and protein columns you have already tried probably allow the greatest degree of control via the mobile phase).

If this were my problem, I would contact the major vendors and ask for their recommendations. Three sources that come to mind are:
- Astec (for cyclodextrin or antibiotic columns)
- Daicel / Chiral Technologies (for derivatized cellulose columns)
- Regis Technologies (for Pirkle columns)

You can find their web sites in the FAQ (drill down to the "where to obtain additional information" section).
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
Thanks for your help!
I also think I need to try other chiral columns.

Not having done any chiral separations myself, I wonder wether there should not be a relatively narrow range of effective retention.
So, bluer,
did your mobile phase span a very wide range of retention or were all of them near the dead time, or all with very high retention?

Those three chiral columns I have tried are all reverse phase columns. I just use the mixtures of water and alcohols as mobile phases. The retention time will be from about 5 to 20 minutes for my samples. But none of those elute conditions can get separated peaks in chromatograms for the enantiomers.

Hans, your point is a good one. In one sense, chiral separations are no different from any other LC separations. The dominant retention mechanism is (usually) either normal- or reversed-phase, and retention is controlled in the usual manner. The enantioselectivity is a "secondary" effect that allows one enantiomer to interact more strongly with the stationary phase than the other.

The retention requirement is the same as for anything else: you want k' values between about 2 and 10.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374

Which probably means that bluer might get a book to see wether there are some more ideas on this, and if he can´t get his columns to work get other ones or try derivatization (making diastereomers)?
7 posts Page 1 of 1

Who is online

In total there are 14 users online :: 1 registered, 0 hidden and 13 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: John Guajardo and 13 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry