SHIMADZU GCMS QP 2010 ULTRA mass calibration error

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

3 posts Page 1 of 1
Hello,

we have a SHIMADZU GCMS QP 2010 ULTRA.
Our problem is a mass calibration error. It appeared first after the change of the
configuration and exchange of a filament. Replacing the two filaments gave errors
"Filament consumed" and "Trap current insufficient.
We found out that the ion source might need a cleaning.
Now we cleaned the ion source. Both filaments were OK and after
cleaning they worked. The two filament errors disappeared,
but now we get again the mass calibration error.
So autotuning doesn't work.
The tension of lense 1 goes up to about 3, 4 or even 5 Volt.
At the end it stays at about 3. Off course,
no tuning signal is displayed.
It seems enough standard PFTBA is still present.
According to SHIMADZU only the ion source
can be the problem.
Putting a new filament seems so easy, but problems appeared
after doing this. What can go wrong in doing this???

What might be the problem?

Thanks

Thomas
without knowing what happened, the mass calibration error and the high lens tensions disappeared
after two days more or less . Even if I improved the vacuum a little bit this does not seem to be reason for the
problem because a few days late with a vacuum less good we got no error of this kind any more.
What ever happened... the error has gone.

Thomas
I have the same problem. I change the filament 2 to 1 and it happend the same problem. This message appear after ion suorce cleaning.
3 posts Page 1 of 1

Who is online

In total there are 23 users online :: 1 registered, 0 hidden and 22 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot] and 22 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry