set column head pressure cant be achieved - GC-3800 VARIAN

Discussions about GC and other "gas phase" separation techniques.

2 posts Page 1 of 1
please,help me, Varian GC cp-3800 , "not-ready'status always appear, because the set column head pressure cant be achieved by the instrument .
I checked for leaks and changed septum and liner,nothing happened. the same problem still exists.
what do you suggest
Welcome to the forum.

How did you check for leaks ?

Is there pressure in the carrier gas supply line to the GC, and is it high enough ?

What has changed since the instrument was last working correctly, what maintenance was done, was a column changed etc etc ?

What are the pressure and flow settings, and the settings for column dimensions in the method ? do the column dimensions match the column that is actually in the GC ?

Peter
Peter Apps
2 posts Page 1 of 1

Who is online

In total there are 63 users online :: 3 registered, 0 hidden and 60 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Baidu [Spider], Bing [Bot], Google [Bot] and 60 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry