Dear friends

I am working with determination of mycophenolic acid in plasma samples by LC-MSMS.
I developed a method to exctract and optimized the MS.

I have a variation of ±10% between injections (n=7-15) of the same vial no matter if I test pure standard or extracted sample.

mobile phase> acetonitrile-acetic acid 1mM (8_2) isopropanol 5%
ion source temperature> 450°C
flow 0.8ml_min
split 1_1

Does anybody has an idea about the reason of this variation ? I am not working with my limit of quantitation value in other words, it is not a matter of low concentration.

Hope to hear from you soon......

Hi Maristela

When you say you have a variation of 10% do you mean the absolute peak areas or the ratio to the internal standard. What internal standard are you using? mycophenolate can be difficult if you use the wrong one.
Geoff

[size=75][color=blue]There's a saying amongst prospectors 'Go out looking for one thing and that's all you'll ever find'[/color][/size]
2 posts Page 1 of 1

Who is online

In total there are 18 users online :: 0 registered, 0 hidden and 18 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 18 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry