Chromeleon 6.80: Lock workstation possible?

Discussions about chromatography data systems, LIMS, controllers, computer issues and related topics.

7 posts Page 1 of 1
Dears

We are using chromeleon 6.80 in a regulated environment with user management and audit trail enabled. Chromeleon can be locked by File -> Lock client. This is good, but not very convenient. Is there an option to automatically lock chromeleon after inactivity for a certain time?
I don't want to lock the workstation but only the chromeleon session, other users should still be allowed to start a second session for data evaluation.

Thanks a lot,
Jörg
Hi Jörg

Is this what you are looking for? Check out user data policies in the user management.

Image

Best regards
Stefan Brand
Thermo Fisher Scientific
Yes, thank you. This is exactly what I was looking for!
Dears

We are using chromeleon 6.80 in a regulated environment with user management and audit trail enabled. Chromeleon can be locked by File -> Lock client. This is good, but not very convenient. Is there an option to automatically lock chromeleon after inactivity for a certain time?
I don't want to lock the workstation but only the chromeleon session, other users should still be allowed to start a second session for data evaluation.

Thanks a lot,
Jörg
BTW, if you are in a regulated environment, sharing Windows login should be strictly prohibited regardless of what access control you have for the application.

BTW, if you are in a regulated environment, sharing Windows login should be strictly prohibited regardless of what access control you have for the application.
I don't think so. Our IT has created a special account for all analytical instruments with limited access to the network drives. So there is no danger of creating files in someone else's name (which is the main reason of having user accounts at all, I guess).

Jörg
We have the same solution in our (regulated) lab.

Shared Windows-login and individual login to Chromeleon. We are thinking of removing the Windows-login completely since it makes no sense. The computers have no internet access or access to any servers (except the Chromeleon server)

Would this be questioned by the authorities? I am not an IT person, is there another way to solve this?

We have two HPLCs / computer so two persons must be able to use each computer
We have the same solution in our (regulated) lab.

Shared Windows-login and individual login to Chromeleon. We are thinking of removing the Windows-login completely since it makes no sense. The computers have no internet access or access to any servers (except the Chromeleon server)

Would this be questioned by the authorities? I am not an IT person, is there another way to solve this?

We have two HPLCs / computer so two persons must be able to use each computer
It will be questioned by the regulators. But, every company is different. It is a good idea to have documented justification (risk assessment, etc) on the reason you are not enforcing person specific access control for that workstation if it is used for GxP activities. I think the workstation including the OS is part of the computerized system. For us, we found that the risk of non-complaint outweigh the convenience.

I am not familiar with the Chromeleon CDS, is it technologically impossible to open a second session on a separate Windows account while the data acquisition session is running under a locked Windows account?

Here are few of the risks for my company;

1) Operator often forgot to manually lock the CDS session. If two similar looking CDS sessions open side by side, operators may access the wrong session (un)intentionally before the automatic lock is triggered. Person specific Windows login will not eliminate this problem, but will significantly reduce the risk.

2) Windows Auditing Logs in the workstation is not attributable. We retain Windows Audit log to reconstruct events, detect and investigate problems. With shared login, the audit trails are not attributable.

3) A person specific Windows login reduces the risk of unauthorized access by adding a second layer of protection.
7 posts Page 1 of 1

Who is online

In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry