Experiencing an issue where RT's have been shifting (>5 shifts)...RT of one injection may be 5 minutes and the next 9, then 6, and so on. Seeing this issue across methods. This is on a purged packed inlet and also have noticed the pressure actuals are higher than the pressure setpoint. Any other ideas that would cause this issue aside from EPC or malfunctioning oven (run is isothermal). Any thoughts/suggestions appreciated.
Hello

I'd recommend to do leak test for PP inlet to see if there is any leak.
I would also check inlet consumables/parts (perhaps replace): septa, liner, viton o-ring and 1/4" Vespel(graphite/vespel) ferrule.

Regards

Tomasz Kubowicz
Thanks, yep - the PP inlet nut was a tad loose and found a pretty good leak there. Upon inspection O-ring looked ok and after tightening the nut the flow actuals and setpoints now agree.
3 posts Page 1 of 1

Who is online

In total there are 37 users online :: 0 registered, 0 hidden and 37 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 37 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry