I have just spent a very frustrating week trying to find the source of some very annoying 'hump-like' ghost peaks in my fatty acid profiles.
I am using an Agilent 6890 GC machine with an autosampler. I am analysing EDTA plasma for fatty acids using direct transesterification preparation and then a 30min split injection (1 or 5ul) GC run starting at 160 degrees for 10 minutes, then 200 degrees, then baking off at the end for 5 minutes at 240 degrees C. The matrix is a mixture of the lipid fraction of human plasma, toluene and methanol.
I'm a bit of a newcomer to GC and chromatography in general but my colleague in the dept is a biochemist specialising in separation science and he is totally stumped too. I would really appreciate some help if anyone has any ideas.
Problem Description:
I can run a plasma sample through once and it is fine. The column is new (I forget the model I'm afraid) - it gives cracking peak resolution and flat-as-a-pancake baselines. However, if I run the same plasma sample through again the second run shows a slightly asymmetrical hump (gentle slope followed by sharp fall back to baseline - about 25 pA in size lasting approx 1min) at about 10-11 minutes, though it can appear later than this. The rest of the chromatogram is perfect.
If I put a blank hexane or methanol after the sample, I get a flat line but with the same ghost peak at 10-11 minutes - the next plasma sample will then be fine. However, I obviously can't be putting a rinse after every sample!
It is almost like each sample is contaminated by the previous one. I have left the GC running over the weekend on a run where each plasma sample is followed by both a methanol and then a hexane rinse - nearly 50hrs runtime for about 30 plasma samples which is pretty OTT but I wanted to get some results out of the batch that I had prepared and they were already getting a bit old. At least the results might tell me more about the source of contamination when I go in tomorrow morning.
Attempted Fixes So Far:
We have (a) changed the water filter (dessicated beads were 'full' so refilled with fresh beads), (b) changed the liner (goose-neck) twice - having found black particles in the old liners though not sure what they were or where they came from, (c) cleaned the inlet port (not sure about the terminology - basically the bit below the autosampler) with methanol, acetone and hexane, and (d) taken a few inches off the column at the injection end. We have also changed the source of the distilled water for the H2 and changed the air canister, though only because it was empty. We have also changed the septum.
I have made new batches of internal standard with totally clean glassware - all reagents are new and unlikely to be contaminated. The problem remains though.
If it is always the second run that contains the ghost hump, I am assuming that this is 'carryover', which I read is caused by overdosing the column with sample. Yet the phenomenon remains even when injecting only 1ul of sample (split from 5ul). Surely this is not too much?
If anyone can help, please please help as I am about to seriously lose it and throw the GC out of the window.
Regards
Simon