Page 1 of 2
VOC Dilution Factor
Posted: Tue Dec 06, 2016 2:04 pm
by ucanmoruk
Hi!
I am working on developing a method for quant some volatile organic compounds on GCMS. My way is like that;
For example; benzen limit is 10 ppm. I prepare 2,5 - 5 - 10 - 25 - 50 ppm calibration point.
We taking organic compounds with a activated charcoal tube and its have 150 mg activated carbon in it.
I break the tube and take the activated charcoal on a tube and add 5 mL of methanol (my solvent that i am using for all method). Mix them for 30 min. After 30 min. i take 1 mL to gc vial from 5 mL solution and give the GCMS to read.
Then when i quant the sample is my dilution factor 5/150 or 5/1 ?
Re: VOC Dilution Factor
Posted: Tue Dec 06, 2016 3:41 pm
by HPLC chemist
Learn your math! The answer is;
(Conc. of sample in ug/mL) x (5 mL/1 mL) / (0.150 g / 1000 g per Kg)
Thus, the answer is in ug/Kg
Posted: Tue Dec 06, 2016 5:32 pm
by Steve Reimer
An alternative is to report it ug/tube assuming you extract the all the charcoal in the tube. Then you need to know how much sample the tube represents to get to ppm.
Posted: Tue Dec 06, 2016 7:51 pm
by James_Ball
As Steve said, if you are looking for concentration in a air sample pulled through the charcoal tube, then you find the mg/tube and the volume of air which may be 50 cubic meters. If you had 10mg per tube then you would have 10mg/50cubic meters or 0.2mg/cubic meter.
If it was a water sample passed though the charcoal then you use the volume of water.
Posted: Tue Dec 06, 2016 10:26 pm
by dblux_
You absolutely must determine desorption rates for all your analytes and include them in calculations if they substantially differ from 100%.
In fact adsorption rate on charcoal from air stream should be investigated as well.
PS - your posts better fit to Student Projects section of this forum
Posted: Fri Dec 09, 2016 8:53 pm
by ucanmoruk
I'm so sory about the simple and absurd question for experts like you and the other forum members.
But i have some troubles on preparing VOC calibration before the working sample on tube. I don't have VOC standart mix and therefore i am preparing standarts by myself.
Look. I am looking for benzene, ethanol and isoamylalcohol. The limits in order for that compounds 0.038, 22.8 and 4.32 mg/m3 for 0.012 m3 activated charcoal tube.
My solvent is methanol. If i take 1 ml ethanol in 10 ml methanol its equal to 78,81 mg/m3 (from d=m/v)
take 1 ul in 10 ml meth = 0,088 mg/m3 benzene
take 200 ul in 10 ml meth = 16,184 mg/m3 isoamyl alcohol
The 10 ml mix solution my 5th point of the calibration. Then i dilute the mix solution 1:2:2:2.5:2 rate. Thats my calibration points.
After that; i take the tube sample, break it into 5 ml methanol. Shake it for about 30 minutes. Take 1 ml to gc vial and give it to gcms.
What is my wrong or missing point?
Posted: Fri Dec 09, 2016 10:30 pm
by dblux_
You need to clarify. So far I guess you are going to determine benzene, ethanol and isoamyl alcohol concentration in air. What are these limits ? Are they max concentrations expected in air ? Are you sure that the volume of charcoal in the tube is 0,012 m3 (it's 12 L)? Typical charcoal tube has 2 layers of charcoal with total mass 150 mg.
If you dilute 1 mL of ethanol in methanol and fill up with diluent to 10 mL then ethanol concentration would be 78900 g/m3.
Before we go any further we have to clarify this.
Posted: Sun Dec 11, 2016 1:08 pm
by ucanmoruk
Benzene limit is 0.038 mg/m3
İsoamyl alcohol limit is 4.32 mg/m3
Ethanol limit is 22.8 mg/m3
The limits are for 0.012 m3 air. Because we are collecting to charcoal tube about 0,012 m3 (it's 12 L) air. And yes charcoal tube has 2 layers of charcoal with total mass 150 mg.
Yes that's true. Ethanol concentration is 78900 g/m3 so 78.9 mg/m3. My reporting unit is mg/m3
Posted: Sun Dec 11, 2016 1:42 pm
by dblux_
78900 g/m3 equals to 78900000 mg/m3. Sorry, these are really the basics.
Posted: Sun Dec 11, 2016 2:15 pm
by Peter Apps
Posted: Sun Dec 11, 2016 8:55 pm
by ucanmoruk
Sorry, thats my fault. I am not fine :)
Ethanol d: 789 kg/m3
Solvent v: 10 ml
I take from ethanol 1 ml = 1000 ul
1 m3 = 1000 L
789 kg/m3 = 789 mg/ml
(1 ml ethanol * 789 mg/ml) / 10 ml solvent = 78.9 mg
If i take 1 ml ethanol to 10 ml solvent, i have 78.9 mg ethanol on this solution
Posted: Sun Dec 11, 2016 10:20 pm
by dblux_
Solvent v: 10 ml
1 m3 = 1000 L
Not exactly. If you add 1 mL of EtOH to 10 mL of MetOH you will get ca. 11 mL of solution not 10 mL. Probably you are going to add 1 mL of ethanol to MetOH and dilute it to 10 mL in volumetric flask.
Nevertheless, the concentration of solution will be 78.9 mg/mL.
PS - when you handle small amounts of substances it's better to use mL and mg instead of m3 and kg.
Posted: Mon Dec 12, 2016 6:35 am
by Peter Apps
Solvent v: 10 ml
1 m3 = 1000 L
[/quote]
No - you have 789 mg of ethanol in this solution. You put in 1 ml of ethanol, 1 ml of ethanol has a mass of 789 mg.
You seem to be confused about mass and concentration; a mass of 789 mg in 10 ml of solution gives a concentration of 78.9 mg/ml.
Posted: Mon Dec 12, 2016 8:36 am
by dblux_
Peter, it would be very kind if you modify quotation in your post in such a way that you ommit the words "dblux_ wrote:".
I simply don't want to be connectet in any way with such ignorance with concentration calculations.
Posted: Mon Dec 12, 2016 10:12 am
by Peter Apps
Done