Discussions about GC and other "gas phase" separation techniques.

2 posts Page 1 of 1
we would like to know the TCD conditions (column temp, injector temp, carrier gas flow rate, auxillary temp, detector temp) for separation of CO,CO2,CH4,N2,C3H6 using H2 as a carrier gas. we are using molecular sieve 13x (60/80 mesh, 30 ft) for this. plz recommend us the required TCD conditions for that.
aslo let me know the retention time for each gases.
dinesh_mantri1@yahoo.com

tcd

"CO,CO2,CH4,N2,C3H6 using H2 as a carrier gas. we are using molecular sieve 13x (60/80 mesh, 30 ft) "

Your elution order is N2, C1, CO, CO2 for the fixed gases for columns activated at temperatures above 330°C. Columns activated at lower temperatures can have CO eluting with methane or even before methane. For propane I have no data but I would suppose it elutes early in the chromatogram perhaps even with the CO peak.

Your detector can be at any temperature but I would recommend 100°C or higher.

injector and column temperatures should be at ambient or above.

Flow rates for 4mm or 2.1mm ID column should be 20cc/min or higher, and for 0.75mm or 1.0mm ID columns 5cc/min or higher.
2 posts Page 1 of 1

Who is online

In total there are 13 users online :: 0 registered, 0 hidden and 13 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 13 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry