Posted: Thu Dec 01, 2016 4:04 am
by pstaunton
hi,
it seems that everytime I run standards on my headspace theres always one vial that seems to not pick up anything. there should be two peaks coming out at 12.5 and 13min. this is frustrating because this is usually a system suitability vial which will obviously invalidate my run.
I use 10ml amber screw top vials and I always make sure there nice and tight. 6ul of standard is placed into 3ml of sample.
any suggestions to help/stop this occurrence. im not going to talk about method parameters as they cant be changed anyway and it is an isolated incident.
Re: headspace vial duds
Posted: Thu Dec 01, 2016 6:42 am
by Peter Apps
Which is it;
"everytime I run standards"
or
"it is an isolated incident".
What hardware are you using ?
Peter
Re: headspace vial duds
Posted: Thu Dec 01, 2016 6:45 am
by dblux_
Are there any regularities? For instance is it always 1st or 2nd vial?
Finding regularities is a good starting point.
Re: headspace vial duds
Posted: Thu Dec 01, 2016 6:50 am
by pstaunton
it seems more often than not it will happen every time to one vial anytime I perform an analysis.
so yes I would say its everytime I run standards for a new analysis run.
headspace sampler is agilent G1888 with agilent 7890A GC.
if theres any parameters that you require to know obviously I will provide
thanks
Re: headspace vial duds
Posted: Thu Dec 01, 2016 7:06 am
by tkubowicz
Hello
If you don't see peaks at all I would say there is nothing going to GC from HS sampler.
In Agilent G1888 you have two small valves: pressurization valve and vent valve - you should check them first. It is not difficult to diagnose which one (or both) is not working however if you've never done it it could be tricky.
If valve(s) are not working properly (small leak) random problems like yours can happen.
Regards
Tomasz Kubowicz
Re: headspace vial duds
Posted: Thu Dec 01, 2016 11:54 am
by Peter Apps
Let's try to be clear what is going on;
You make up a standard vial by diluting 6ul of standard into 3ml of sample, and when you run it there are no peaks ?
There are never any peaks from standards made up like this. ?
Is this standard vial always at the same place in the sequence ?
Do you make up standards with higher concentrations than this, and do you see peaks when you run them ?
It is a waste of time getting details on methods and hardware a little bit at a time - so go ahead and tell us everything, including what is on the standards and its concentration.
Peter
Re: headspace vial duds
Posted: Wed Dec 07, 2016 2:56 am
by pstaunton
validated method mr apps. below parameters cannot be changed. no, the vial that gives no response seems to be quite random.
I just ran another sequence and my DSS vial this time decided to show no peaks. concentrations for the two peaks in the DSS was 0.5ppm.
the working standards are around 19ppm. stocks are made up in a mixture of water and dimethylacetamide. sometimes the ws shows no peaks.
vial pressure : 16psi
vial oven temp:80C
loop temp:130C
transfer line temp:140C
agitation: high
vial equilibration:20min
vial pressurization:0.4min
loop fill:0.2min
loop equilibration:0.2min
column flow: 5ml/min
split:1:1
to reiterate it happens to one vial out of every 30 vials or so and it can be anywhere in the sequence (dss, system suitability, bracketing std).
Re: headspace vial duds
Posted: Wed Dec 07, 2016 4:37 am
by Peter Apps
If you are spiking vials one at a time then human error has to be a possibility. Make up in bulk enough of a high standard to charge a full sequence of vials, and then divide it between the vials. Then you know you have analyte in every vial. Run.
Or retrieve the dud vial, put a new cap on it, and run it again.
A leaky cap is another possibility - but I would expect at least some peaks.
Do you ever get chromatograms with no peaks (no peaks at all, not just no peaks for the analytes) when you run samples ?
Peter
Re: headspace vial duds
Posted: Wed Dec 07, 2016 6:53 am
by dblux_
...
column flow: 5ml/min
split:1:1
...
Just curious, what is your column ID ?
Re: headspace vial duds
Posted: Wed Dec 07, 2016 8:11 pm
by James_Ball
Do you inspect the rim of each vial. We recently had some 40ml VOA vials where a few per case would have chips in the rim, which would not allow the septa to seal when doing headspace in the vial. Very small chips but enough to interfere with the seal.
Re: headspace vial duds
Posted: Wed Dec 14, 2016 6:13 am
by pstaunton
Hi dblux,
column using is db-5, 30M x 0.53mm i.d.
thanks james ball I will have a look closer at each vial I prep.
Re: headspace vial duds
Posted: Wed Dec 14, 2016 7:38 am
by Peter Apps
If the problem is only with standards and not with samples then it is unlikely to be anything to do with the vials. Unless peak-free chromatograms have only happened a few times.
Peter