Tetrabutylammonium cleanup [August 3, 2004]
Posted: Sun Aug 29, 2004 6:19 pm
by admin
By Bruce Girton on Tuesday, August 3, 2004 - 03:21 pm:
If anyone could offer any insights on the most efficient way to wash tetrabutylammonium out of a C18 column, I would appreciate it.
By A.Mouse on Tuesday, August 3, 2004 - 06:47 pm:
This appears to be a very difficult task, not worth the solution. Column manufacturers recommend not to do anything, and just to buy a new column.
I can not guarantee any results, because until now I listened to the advice I got. But here is some thinking:
Go to an acidic mobile phase, maybe with TFA or formaic acid. This should neutralize the silanols. Then go to a high organic concentration (at the same concentration of acid that you are using). Maybe 80% organic is OK, Now wash the column for a while.
How will you know that you got rid of the TBA? MS?
By Jimmy on Sunday, August 15, 2004 - 08:14 pm:
I would use 0.1% TFA, 70% methanol and 30% ammonium phosphate buffer, 25 mM, pH 2.5. Do check solubility before washing.