Hello all,

I have a simple question. We have a 5890 Series II GC at my work, and I need to change out our column. Unfortunately, the instruction manual was lost long ago (before I ever started here) so I don't quite know the adjustments (ie how much the column should protrude from the ferrule, etc). Does anyone know where I could get an instruction manual?

Or even better, give me the basic necessities for installing a megabore (DB-1) column?

I appreciate your help!

-Robin

I'd go with 4 mm column protruding at the inlet end; at the detector end, push the column up until it hits the FID (a few inches), then pull back 1 or 2 mm and tighten. Agilent has a video showing column installation http://www.chem.agilent.com/cag/cabu/howtovideo.htm and check out http://www.chem.agilent.com/cpdocs/a158 ... N=99646943
Agilent has a good tech support online
2 posts Page 1 of 1

Who is online

In total there are 22 users online :: 2 registered, 0 hidden and 20 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Google [Bot], Semrush [Bot] and 20 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry