I am using an Agilent 6890/5973. Because I found N2 in autotuning and also peaks of dimethylpolysiloxane, i decided to change column. Yesterday, I changed pump oil first and the replaced a new DB-5MS column under the diretion of a Postdoc in our lab. He asked me to connect the two ends of the column to the inlet part and the MSD interface first, and then do conditioning. I realized now that this is absolutely wrong. However, it was operated like that. Before the conditioning, the gauge reading is 3.7*10 to -5 torr, close to what I usually have. But this morning (after conditioning), it is almost 9*10 to -5 torr. I tried to autotune, but tune stopped: cannot achieve constant peak widths.
I don't know if this is an air leak problem or the ion sources need to be cleaned. Why the vacuum got worse after the conditioning? What should I do first?
Thanks a lot. I know I did a stupid thing.