- Posts: 107
- Joined: Sat Mar 11, 2006 6:45 pm
We have one validated LC method (isocratic) to determine enantiomeric purity using mobile phase system 97:3 Hexanes: t-butyl alcohol with Chiral AD-H column. The main peak initially comes out at 65 minutes and then gradually drifts past 100 minutes. The run program intially set at 80 minutes, now it is extended to 120+ minutes (sometimes 150 minutes) in order to capture the the main peak (I have total 5 "pricey" chiral columns and they all behave the same). I tried them all on 4 Agilent 1100 series and got the same issue (I also allow at least 12 hours of equilibration). I also observed pressure fluctuation on 4 LCs. As part of investigation, which I strongly believe mobile phase is the culprit, I prepare another mobile phase system 95:5 Hexanes:t-butyl alcohol and it works like a champ. I set up sequence to run with this new mobile phase for 24 hours and the Retention time of the main peak (comes out at 38 minutes)doesn't shift at all (I have tried on 3 LCs and I don't see any pressure fluctuation). Can anyone explain to me why? It seems to me I got unstable mobile phase using less organic modifier (t-butyl alcohol at 3%) compared to 5% t-butyl alcohol. I'm new to this chiral technology and I
do need your help.
Thanks in advance,