pi3832 » Fri Apr 07, 2006 11:56 am
My carrier gas is nearly empty, I’ve never change the cylinder and as we are on Easter holidays there is nobody in the lab who knows how to change it.
To change the carrier all you need is a full cylinder and a 1-1/8" wrench.
Steps:
1 - Make note of the outlet pressure of the regulator.
2 - Back the regulator dial out all the way (turning counter-clock-wise).
3 - Shut off the valve on the cylinder (clockwise when looking from the top).
4 - Loosen the nut connecting the regulator nipple to the cylinder valve. For helium, nitrogen or argon, the nut is a right-hand thread. (Turn the nut counter-clockwise when looking at it from the regulator side.) For hydrogen, the nut is left-hand threaded, so turn it the other way. (Note: left-hand threaded nuts have a notch cut into the corners.)
5 - Screw the metal cylinder cap on the empty cylinder. (If you don't have the cap anymore, use the one from the full cylinder.)
6 - Move the full cylinder into place, and restrain it from falling with a chain or strap.
7 - Connect the regulator to the cylinder valve, and tighten the nut.
8 - Briefly open the cylinder valve.
9 - Briefly loosen the nut holding the nipple, venting the pressure in the nipple, and then re-tighten the nut.
10 - Repeat steps 8 and 9 at least two more times. This purges the air out of the nipple and the inlet of the regulator.
11 - Open the cylinder valve all the way, and check for leaks. (You can use Snoop or somesuch.)
12 - Dial in the regulator's outlet pressure to the level you noted way back in step one.
Seems like a lot of steps, but it shouldn't take you more than five minutes.
In general, you shouldn't let a cylinder of high purity gas get below 200 psig.
If losing carrier is a re-occuring problem, you can add a back-up cylinder to the system. Just T into the supply line, and set the outlet pressure on the regulator of the back-up well below the outlet pressure on the main regulator. Gas will only flow from the back-up when the main cylinder runs out of pressure and the pressure in the line falls below the set pressure on the back-up. The flows on your GC will be all screwy but you won't have to worry about damage to the column.