more on tailing, acetone, non-Wax phases --> my analytes
Posted: Thu Apr 06, 2006 3:02 pm
First, I apologize if this is getting to be redundant for some, but this has been a real challenge for me. And, I still don't know what's going on...
To review:
My analytes were phenolic aldehydes which have been chlorinated by a fungus. i.e. aromatic ring, OH, CHO, OMe, Cl and monomeric.
My samples have been cleaned up by SPE and eluted with acetone, which is a very good solvent for lignin and the like. I have no choice over the solvent.
I started the GC using a ZB-5ms column. First few runs looked OK, but then tailing developed and worsened horribly. Tried 2 other manufacturers columns (Supelco and Varian Factor Four) which also exhibited severe tailing. Got so bad there was almost no discernable peak from the standards.
Switched to VF-200ms column marketed for use with electron-rich compounds like halogenated aldehydes. Still very bad tailing.
Started looking very closely for instrumental sources:
Verified split flow and septum purge, clean inlet, ultra-high purity helium with 2 different in-line high-capacity gas purfiers with color indicators.
Removed all packing from liners. Tried both deactivated glass and Siltek-coated single gooseneck liner. Same result: bad tailing.
Ion trap is clean as a whistle.
Removed pressure pulse from splitless injection.
Did split injection and still observed bad tailing.
Lowered injector temp from 280 to 220. No improvement.
Raised initial column temp from 40 to 140. No improvement.
Clipped head of column, then clipped head & tail.
Using highest quality acetone I know (B&J GC2).
Made up stds in MeCl2 (B&J GC2) and still saw bad tailing.
Made up indiviudal stds in MeCl2 to ensure no interaction between the target compounds in the inlet. Still tailed badly.
Speculation was that the acetone had ruined the column so making the stds up in MeCl2 had no effect because of the previous acetone.
NOTE: Grob Mix still looked same as new: tailing diol & amines; 4-Cl-phenol was OK, nonanal was not OK and showed presence of nonanoic acid. But since this had not changed since the column was new, I assumed the column was still OK.
Injected a series of similar compounds which did not contain all of the functionality of my analytes. Other than the 3,5-t-butyl-4-OH-benzyl alcohol, all showed very nice peak shape and they were in acetone.
Finally, followed suggestion by Peter Apps to try a Wax column. That did the trick!
Ran very long sequence with multiple samples, bracketed by 10-point standard curve, including 5 QCs within the sample sequence and each and every sample/std/QC was followed by at least 1 acetone rinse to ensure no carry-over (to prove source of chlorination). (Chlorinated aldehydes give very pretty pattern of M+, M+2, M+-1, M+-1+2.) Beautiful results, very respectable linearity and sustained response.
Then it occurred to me that while the Wax column worked just fine with the acetone solvent, I had not truly proved that acetone is incompatible with the bonded phase column. i.e. was it the acetone or was my analytes?
The only brand-spanking new column I have is a ZB-1ms.
With utmost care (i.e. cool injector, oven and transfer line; continued flow) I installed the brand new ZB-1ms not connecting it to the MS. Swabbed the injector and installed new siltek single gooseneck liner. After flowing 15 min with high split to purge injector, I conditioned the column 40C for 15 min, to 300C at 5C/min, hold 30 min. I cooled everything again, connected to MS, and conditioned again. I did everything I could to ensure no oxygen saw this column while it was hot.
Then I injected a standard mixture in MeCl2.
On this column that has never seen acetone, my analytes tail and have significantly decreased response. The non-chlorinated internal standard (4-OMe-cinnamaldehyde) does not tail. Whereas, on the Wax column this standard gave approximately same response for all chlorinated analytes and the internal std with all having good peak shape.
So, yesterday I made up another batch of the similars only I used MeCl2. the similars are: 3-OMe-4-OH-benzaldehyde (vanillin), 3,4-OMe-phenyl acetone, 3,5-chlorophenol, 4-OH-3-OMe-benzyl alcohol, 4-OEt-3-OMe-benzaldehyde, 4-OMe-benzyl alcohol, 3,5-OMe-4-OH-acetophenone, and 3,5-t-butyl-4-OH-benzyl alcohol. All, except the last which gave low response, gave excellent peak shape.
---> The similars which have some, but not all, of the analyte functionality do NOT tail. Whereas the analytes DO tail - even in MeCl2 - and using a brand new column.
---> Now I'm not so sure the culprit is the acetone.
The culprit may be the analytes themselves.
QUESTIONS: Have you ever seen something similar? Any other thoughts?
To review: