High pressure before every sample batch
Posted: Thu Nov 24, 2016 11:04 am
Dear folks,
we have some trouble with our new HPLC system (Agilent G1312B 1260 Bin Pump). We run an LC method with a start gradient of 60% A (1.5% acetic acid, 1 mM NH4OAc, water) and 40% B (1.5% acetic acid, 1 mM NH4OAc, methanol) at 0.3 mL/min. The pressure at 60% A and 40% B is usually around 470 bar. The sample batch always works fine with similar pressures before the injections. After every batch, the LC is automatically turned off. Here is the problem: When starting the LC again the next morning, the pressure goes immediately high up to 600 bar. After rinsing for 1h, pressure is still the same, it doesn't normalize. To solve the problem, we turn off the pump for a while and after turning on again, the pressure is usually back to normal. Sometimes we have to turn off the pump again to solve the problem.
We tried to run the LC after the sample batch has finished at a low flow rate (0.05 mL/min) instead of 0 mL/min, and the first time it worked fine, normal pressure when adjusting to 0.3 mL/min. But the next morning, it didn't work, we had 600 bar again. Any comments and recommendations?
we have some trouble with our new HPLC system (Agilent G1312B 1260 Bin Pump). We run an LC method with a start gradient of 60% A (1.5% acetic acid, 1 mM NH4OAc, water) and 40% B (1.5% acetic acid, 1 mM NH4OAc, methanol) at 0.3 mL/min. The pressure at 60% A and 40% B is usually around 470 bar. The sample batch always works fine with similar pressures before the injections. After every batch, the LC is automatically turned off. Here is the problem: When starting the LC again the next morning, the pressure goes immediately high up to 600 bar. After rinsing for 1h, pressure is still the same, it doesn't normalize. To solve the problem, we turn off the pump for a while and after turning on again, the pressure is usually back to normal. Sometimes we have to turn off the pump again to solve the problem.
We tried to run the LC after the sample batch has finished at a low flow rate (0.05 mL/min) instead of 0 mL/min, and the first time it worked fine, normal pressure when adjusting to 0.3 mL/min. But the next morning, it didn't work, we had 600 bar again. Any comments and recommendations?