Posted: Thu Nov 24, 2016 3:52 am
I had been running a 5971 with good results for a couple of months, when over the course of a few days the high mass sensitivity dropped. Oddly, the mid mass sensitivity seemed to increase. I am running hydrogen carrier gas, and my vacuum is better now than it has ever been at 2.0x10-5.
So far, I've cleaned the source (it appeared perfectly clean), changed the filaments (they looked fine), changed the every board except the power supply (re-checking results in between). I checked the power supply voltages, and they are fine. Rechecked my column flow (but I get the same results even with the transfer line removed and capped).
I'm pulling my hair out. I finally changed the detector for a new one, but it hasn't made a difference. The odd thing, is that I cannot even raise the EM voltage above 1400 with the new detector, because when I do that, it throws the ratios way off and the auto tune then reduces the EM voltage to bring everything in line.
Perhaps there is some calibration I'm not aware of? Any suggestions would be greatly appreciated.
So far, I've cleaned the source (it appeared perfectly clean), changed the filaments (they looked fine), changed the every board except the power supply (re-checking results in between). I checked the power supply voltages, and they are fine. Rechecked my column flow (but I get the same results even with the transfer line removed and capped).
I'm pulling my hair out. I finally changed the detector for a new one, but it hasn't made a difference. The odd thing, is that I cannot even raise the EM voltage above 1400 with the new detector, because when I do that, it throws the ratios way off and the auto tune then reduces the EM voltage to bring everything in line.
Perhaps there is some calibration I'm not aware of? Any suggestions would be greatly appreciated.