by
KC » Tue Apr 04, 2006 12:39 am
I would change the septum and the liner, probably clip the inlet end of the column 3-6". Then after putting it all back togeather set inj temp about 40-60 C higher than your method (but not above detector temp) and cook the column overnight at about 310 if your max temp is 325. Wait for somebody else to confirm what I have said cause I don't want to be held responsible but truthfully, thats what I would do. May even shoot a bunch of solvent blanks like acetone but not sure if that would help much if you have changed septum and liner. As far as that split ratio and flow I am not sure, sounds a little high but am not use to running much other than splittless. 40C ramp in temp gradient should not be a problem, have used it in methods many times (would use more at times if software would allow).