Hello. I have some trouble with my unit Varian 3400cx. It's display "Fault 13" and " Fault 37". I don't have manual for my unit. Anybody can tell my what this is errors about? Thanks.
-
- Posts: 131
- Joined: Wed Mar 23, 2011 7:06 pm
I'm at home now. I will look the manual next monday
G.Serena
serena.guillermo@gmail.com
serena.guillermo@gmail.com
-
- Posts: 131
- Joined: Wed Mar 23, 2011 7:06 pm
Next monday I will have a look yo the manual.
Regards
Regards
G.Serena
serena.guillermo@gmail.com
serena.guillermo@gmail.com
-
- Posts: 56
- Joined: Mon Apr 05, 2010 6:58 pm
Hello,
I can send You the manual in pdf (4,3MB) via email.
13 Power fail occurred (latched)
AC power to the instrument was interrupted, either by a power failure or by turning off the main power switch
37 Power line ground fault (latched)
The polarization of the power line connection is reversed, or the ground connection through the power cord is missing. Normal chromatographic functions are not affected, but the risk of electrical shock is increased for certain maintenance operations and in case of some component failures.
So, first just try to turn 180 the main power plug.
Regards,
I can send You the manual in pdf (4,3MB) via email.
13 Power fail occurred (latched)
AC power to the instrument was interrupted, either by a power failure or by turning off the main power switch
37 Power line ground fault (latched)
The polarization of the power line connection is reversed, or the ground connection through the power cord is missing. Normal chromatographic functions are not affected, but the risk of electrical shock is increased for certain maintenance operations and in case of some component failures.
So, first just try to turn 180 the main power plug.
Regards,
-
- Posts: 2
- Joined: Sat Nov 12, 2016 4:30 pm
Please can you send me (patsay@mail.ru) the manual for Varian 3400 ?Hello,
I can send You the manual in pdf (4,3MB) via email.
13 Power fail occurred (latched)
AC power to the instrument was interrupted, either by a power failure or by turning off the main power switch
37 Power line ground fault (latched)
The polarization of the power line connection is reversed, or the ground connection through the power cord is missing. Normal chromatographic functions are not affected, but the risk of electrical shock is increased for certain maintenance operations and in case of some component failures.
So, first just try to turn 180 the main power plug.
Regards,
5 posts
Page 1 of 1
Who is online
In total there are 27 users online :: 0 registered, 0 hidden and 27 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 27 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 27 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science