Does contamination affect vacuum?
Posted: Sat Apr 01, 2006 6:33 pm
I got a low vacuum after changed the column (DB-5) on 6890/5973.
I removed the column from the injection port, and capped off. The system pumped down very quickly and reached a very good vacuum. So the leak is isolated to GC. I replaced the liner O-ring, septum, and installed the column back to the injection port, then the vacuum went back to ~9*10(-5) torr again.
I did manual tune and find many other ions except 69 and 219. However 28:32 is about 2:1, so it is not an air leak. Do you think there is a contamination of the gas? Does the contamination affect the vacuum? The trap is 2-year old now.
Thanks
I removed the column from the injection port, and capped off. The system pumped down very quickly and reached a very good vacuum. So the leak is isolated to GC. I replaced the liner O-ring, septum, and installed the column back to the injection port, then the vacuum went back to ~9*10(-5) torr again.