I am having a little trouble with my TD-GC-MS at present. The system I am currently using is an Agilent 7890 B Gas Chromatograph coupled to an Agilent 5977 A Mass Spectrometer. I am currently running a dual inlet/column system with the rear inlet and column connected to a Markes UNITY2 thermodesorption unit for analysis of VOCs trapped on sorbent tubes, while the front inlet and column are reserved for liquid injections. Both of the columns are connected to a deans switch that outputs into a single MSD transfer line.
The issues I am having appear when trying to run an MS tune, where it states that I have an air leak. I have two methods (one for liquid injections on the front column and one for TD injections on the rear column) that switch the column pressures depending on the method used. When the front column is in use the front column pressure is set to 10 PSI and the rear column pressure is set to 4 PSI. The reverse is true when the rear column is in use. I have attached the tune files below:
Front Column Tune

Back Column Tune

I'm not sure why I am failing the air/water check when using the rear column but not the front column.
Any help would be appreciated.
Cheers,
Joe