						adsorption of amlodipine besylate on filter after dissolution ask if any another method to exclude filtration after dissolution with put onto consideration centrifuge perform the same problem on dissolution results can i take sample as it is for injection on HPLC ???
					
									formulator
				- 
																									  
- tom jupille
- Site Admin
- 
								
 
- Posts: 4978
- Joined: Wed Aug 11, 2004 4:55 pm
						Several comments here:
1. Have you tried different filter media to find something less adsorptive?
2. Centrifugation should work. However if you find evidence of adsorption on the pellet (i.e. incomplete recovery from a spiked matrix blank), then you really have to revisit your formulation, because that suggests that you will get poor availability in vivo
3. Direct injection into the HPLC may be possible if the volume injected is small enough, but I would *definitely* use an in-line filter and/or a guard cartridge -- and be prepared to change them frequently.
									1. Have you tried different filter media to find something less adsorptive?
2. Centrifugation should work. However if you find evidence of adsorption on the pellet (i.e. incomplete recovery from a spiked matrix blank), then you really have to revisit your formulation, because that suggests that you will get poor availability in vivo
3. Direct injection into the HPLC may be possible if the volume injected is small enough, but I would *definitely* use an in-line filter and/or a guard cartridge -- and be prepared to change them frequently.
-- Tom Jupille 
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
				LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
- 
								
 
- Posts: 2
- Joined: Thu Nov 17, 2016 6:02 am
						centrifuge perform the same problem can i take sample as it is from dissolution vessel to inject it to HPLC direct ???
					
									formulator
				
			3 posts
							Page 1 of 1
					
	Who is online
				In total there are 19 users online :: 0 registered, 0 hidden and 19 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
 
		
Users browsing this forum: No registered users and 19 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 19 guests
Latest Blog Posts from Separation Science
    Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
    
    
    Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
    
    
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science