What is the difference between vespel and graphite ferrule. And can I use graphite as inlet ferrule to MSD column and MSD interface?
-
- Posts: 326
- Joined: Wed Oct 02, 2013 12:20 pm
Hello
There is a lot of materials available on manufacturers websites...
For example see links below:
http://www.bioszeparacio.hu/_user/brows ... errule.pdf
https://www.sepscience.com/Techniques/G ... c-Ferrules
Regards
Tomasz Kubowicz
There is a lot of materials available on manufacturers websites...
For example see links below:
http://www.bioszeparacio.hu/_user/brows ... errule.pdf
https://www.sepscience.com/Techniques/G ... c-Ferrules
Regards
Tomasz Kubowicz
-
- Posts: 5433
- Joined: Thu Oct 13, 2005 2:29 pm
Vespel is hard and graphite is soft.
Mixtures of graphite and vespel have intermediate hardness/softness.
Grahpite ferrules deform very easily and shed particles that conduct electricity - these are bad news in a mass spec, so do not use graphite ferrules on the MS end of the column.
Peter
Mixtures of graphite and vespel have intermediate hardness/softness.
Grahpite ferrules deform very easily and shed particles that conduct electricity - these are bad news in a mass spec, so do not use graphite ferrules on the MS end of the column.
Peter
Peter Apps
-
- Posts: 3477
- Joined: Mon Jan 07, 2013 8:54 pm
As Peter said, you should only use the graphite/vespel mixed or vespel ferrules on the MSD interface, the 100% graphite will leak and can contaminate the ion source. Vespel and graphite/vespel ferrules do shrink when heated, so you will need to retighten the fitting a day or two after installing, or keep a few ferrules in the oven in a beaker so the heating of the oven will pre-shrink them before use.
100% graphite is ok for the column inlet/injector end of the column. If you will be doing maintenance such as trimming the column on a daily basis, then graphite is the best choice since it doesn't adhere to the column and is removed and reinserted easily. Just be certain that you always trim about 1cm of column from the end AFTER you put the ferrule on the column to be sure you remove any pieces of graphite that might scrape off the ferrule as the column passes through it.
100% graphite is ok for the column inlet/injector end of the column. If you will be doing maintenance such as trimming the column on a daily basis, then graphite is the best choice since it doesn't adhere to the column and is removed and reinserted easily. Just be certain that you always trim about 1cm of column from the end AFTER you put the ferrule on the column to be sure you remove any pieces of graphite that might scrape off the ferrule as the column passes through it.
The past is there to guide us into the future, not to dwell in.
-
- Posts: 517
- Joined: Thu Jul 14, 2016 12:08 pm
I can confirm that what the posters above me said is also common practice in our lab.
Graphite or mixed vespel/graphite (=VG) for inlet, and VG for the MSD end. I can add that we prefer to use relatively larger (longer) ferrules for the MSD end, since in our experience air leaks happen more often with the smaller ones.
Graphite or mixed vespel/graphite (=VG) for inlet, and VG for the MSD end. I can add that we prefer to use relatively larger (longer) ferrules for the MSD end, since in our experience air leaks happen more often with the smaller ones.
5 posts
Page 1 of 1
Who is online
In total there are 11 users online :: 0 registered, 0 hidden and 11 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 11 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 11 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science