hello I have a simple question. Is the detector part of the mass spectrometer held under vacuum conditions. Some texts i have read say it is and some say it isnt ! :?:

Depends on the type of mass spectrometer you are interested in.

Well, since what we are doing in MS is analyse ions generated by the ion source, the whole system should be held under vacuum to avoid interferences and allow the ions migrate to the detector.

The ionization chamber might be at lower vacuum - depending also the kind of ion source you have and which kind of system you are interfacing - but the detector is usually in the 10-6 mbar range.

Maybe you are confusing mass spectrometers with ion mobility spectrometers?

see

http://www.chemistry.org/portal/a/c/s/1 ... 245d830100
Sailor

Funny stuff!

I have heard about this technique in lab corridors but I did not knew they are actually in use!
5 posts Page 1 of 1

Who is online

In total there are 15 users online :: 0 registered, 0 hidden and 15 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 15 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry