Dear Friends
I am working with a drug analysis in plasma samples by LC-MS-MS.
I extract 50ul of sample with 1 ml of diethylether-hexane (8:2 v/v). My samples concentration are between 100 ng/ml to 40000 ng/ml.
The mobile phase is acetonitrile-ammonium acetate 2 mM (7:3 v/v), the injection volumn is 10ul and the run time is 3.5 min.
I have a decreased of 20% of the analyte signal on a list of 50 samples injected.
The internal standard did not decreased at the same way. (it is not a deuterated internal standard and I think it does not respond the same way)
After each 50 samples I wash the column with acetonitrile during 40 min and it return to the same response of the beggining.
Is it a column saturation problem ? Is methanol better than acetonitrile to avoid this " accumulation " in the column ???