Waters 2414, error message "LED intensity too low"

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

2 posts Page 1 of 1
Keep getting this error message. Have purged for significant time.

Only running water through the system.

Any suggestions would be greatly appreciated.

Scott
Hi

You should perform "Optimize LED" procedure.
Try to do below:

1.Press "DIAG"
2.Choose "Optimize LED"

Of course make sure you have cell properly flushed with water - you can try flush it backwards (just swap inlet with outlet)

Good luck

Regards

Tomasz Kubowicz