Our laboratory does a good deal of vitamin C analysis and we use a weak anion exhange column using an amine phase column. The column is conditioned with mobile phase of 75% acetonitrile and 25% 0.15M Sodium Acetate Buffer, pH=5.0. This chromatography is similar to carbohydrate analysis (mono- and di-saccharides). We use a calibration from 2.5 ug to 50 ug with good linearity.
People who do this analysis use dithiothreitol as a reducing agent (to keep asocorbic acid from oxidizing to dehyroascorbic acid and then to di-ketogluconic acid). Once oxidized to di-ketogluconic acid, it will not be able to be reduced back to asocorbic acid. So the analysis and results you are achieving might be from oxidation of your standard. In the food industry, meta-phosphoric acid is also used in the extraction solvent (up to 5%) and this does not seem to affect the linearity of standards. Our lab also adds a small amount of EDTA to complex any free metals (mostly iron) to prevent oxidation.
On the other side of this, some labs do a total oxidation of the material and a derivatization and then HPLC (C18, methanol:water combination).
Our lab actually uses a narrow bore (2 mm) column for solvent savings. I hate wasitng all that acetonitrile. Good Luck.