Ascorbic Acid Ion Exchange Oddity

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

5 posts Page 1 of 1
DawnRazor,

You could have an highly retained component in your sample (at least relative to ascorbate and phosphate at pH 6.5) tying up your ion exchange sites until eluted with 500mM buffer. What is in the sample besides ascorbate?
DawnRazor,

This may possibly be avoided at higher ionic strength. What is your sample matrix, just ascorbate and buffer? Surely there is more in the sample than this.

Ascorbate oxidizes rapidly at neutral or alkaline pH, and is catalyzed by transition metal ions. Are you sure your sample is stable?
Mark Tracy
Senior Chemist
Dionex Corp.
Our laboratory does a good deal of vitamin C analysis and we use a weak anion exhange column using an amine phase column. The column is conditioned with mobile phase of 75% acetonitrile and 25% 0.15M Sodium Acetate Buffer, pH=5.0. This chromatography is similar to carbohydrate analysis (mono- and di-saccharides). We use a calibration from 2.5 ug to 50 ug with good linearity.
People who do this analysis use dithiothreitol as a reducing agent (to keep asocorbic acid from oxidizing to dehyroascorbic acid and then to di-ketogluconic acid). Once oxidized to di-ketogluconic acid, it will not be able to be reduced back to asocorbic acid. So the analysis and results you are achieving might be from oxidation of your standard. In the food industry, meta-phosphoric acid is also used in the extraction solvent (up to 5%) and this does not seem to affect the linearity of standards. Our lab also adds a small amount of EDTA to complex any free metals (mostly iron) to prevent oxidation.
On the other side of this, some labs do a total oxidation of the material and a derivatization and then HPLC (C18, methanol:water combination).
Our lab actually uses a narrow bore (2 mm) column for solvent savings. I hate wasitng all that acetonitrile. Good Luck.

Is there a reason, here, for not using mobile phases at lower pH where ascorbic is little ionized and stable against oxidation? (No ion pairing needed)

Incidentally, if one knows somebody with an electron spin resonance machine one may check if there is the characteristic radical present in ones samples. If yes the pH is too high.
5 posts Page 1 of 1

Who is online

In total there are 8 users online :: 1 registered, 0 hidden and 7 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Bing [Bot] and 7 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry