Can someone please explain this to a beginner in HPLC.
I have a standard solution that contains one polar analyte, and that is only diluted with water. I run it in a mobile phase that consists of 80% acetonitrile and 20% water, the detector is a RI detector.
At low concentrations (0.1 to 20 mg/mL) of the standard I get 2 resolved peaks, one for the analyte and one for the solvent (?) =water.
At higher concentrations (20 mg/mL and higher) of the standard, the analyte peak migrates into the solvent peak, which results in one very distorted positive peak, with a negative valley on its right side. Sometimes the result is one large distorted peak and one smaller peak with a valley between them.
My questions:
Why is the solvent peak positive at low concentrations? Shouldn´t it be negative since water has a lower RI value than acetonitrile?
And why do I get a negative peak and distorted positive peaks at higher concentrations? Is it due to column overload or what?
I am very thankful for any suggestions.