Chemstation software for Agilent MSD 5975B and GC 6890N: D.0300611. We have been having issues with the program shutting down in the middle of a run. No error codes but machines are still running (they say acquiring sample but they are not because the software shut down) Any ideas on a starring point. We've tried shutting down and restarting but that didn't help.
When we start the software back up the chemstation log book says: startup initiated after unexpected shutdown. This appears to be happening around the same time each night.
We've had similar things happen, sometimes with error message and sometimes not. One simple thing that has often helped has been to open up the MSD Configuration software and just reconfigure the instrument.
did you disable Windows auto lock timer and screen saver?
4 posts Page 1 of 1

Who is online

In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry