Empower3 custom field

Discussions about chromatography data systems, LIMS, controllers, computer issues and related topics.

3 posts Page 1 of 1
Does anyone know how to set up the custom field to calculate the linear curve ?

Thanks,
ML
In processing method's components tab, name your standard peak(s), add retention times, choose linear or linear through zero fit type.
In sample set method, set function to "Inject standards" and either "normal" or "do not process or report" for Processing for each standard (whose Labels should start with "S"). If you choose the latter, include a Calibrate line for standards at the end of the sample set method, followed by a Quantitate line for samples (unless you go with "Normal" for them).
Collect and process your data.

There is absolutely no need of a custom field for any of this.
Thanks,
DR
Image
Thank you for your help!

Best Regarda,
Maily
3 posts Page 1 of 1

Who is online

In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry