Guard cell clogged, electrochemical detector

Posted: Tue Oct 18, 2016 12:45 pm
by Magestroz
Hi,
The guard cell of my electrochemical detection-HPLC is clogged. Maybe any of the readers know a maintenance protocol to try to clean it?

The guard cell is a ESA Coulochem 5020 guard cell, and consists of a graphite electrode able to support high pressures. You can see here its design:
http://www.spectrotech.com.tw/esa/cell.htm

My mobile phase is an aqueous solution containing 15 % methanol. I have tried 100 % methanol and 100 % acetonitrile to clean it; the backpressure went down a bit but it´s still very high. I'm note sure if it would be a good idea to backflush any of those solvents through the guard cell or to use stronger solvents (methylene chloride, THF)??

M

Re: Guard cell clogged, electrochemical detector

Posted: Wed Oct 19, 2016 7:32 am
by sbrand
Hi
Yes you can backflush the guard cell. If you drop me your email I will send you some technical notes from ESA about Maintaining Cell Performance.
Wishing you success.
Regards

Re: Guard cell clogged, electrochemical detector

Posted: Wed Oct 19, 2016 7:54 am
by Magestroz
Hi,

Those technical notes would be more than welcome!
My e-mail is manugesto at gmail.com

Thank you in advance,

M

Re: Guard cell clogged, electrochemical detector

Posted: Wed Oct 19, 2016 12:54 pm
by sbrand
email sent.
Best regards