ACIDS LIKE TARTARIC ACID ETC.
Posted: Sat Mar 18, 2006 9:40 am
by N.V.V.S.S. RAMAN
Dear Friends,
When I developing methods for salts like fumerate, tartrate, maleate of some drugs, I am getting problem of these salts. They elute at below 2 min and getting merged with blank . To avoid this blank intereference is very difficult. Whatever mobile phase and whatever column may be , the problem is persisted.
Can anybody suggested a solution
raman
Posted: Sat Mar 18, 2006 9:52 am
by zompa
Hi Raman,
I don't know if it could be helpful to you but everytime I analyse such compounds I use the ionic chromatograph... it works!
Posted: Sat Mar 18, 2006 9:56 am
by ym3142
google.
Posted: Sat Mar 18, 2006 10:19 am
by N.V.V.S.S. RAMAN
Thank you Zompa
But in our case, while analysing the drug this salt will also respond at a lower level. So we are asked to address and quantify the same at this level. That is the reason we are struggling.
Raman
Posted: Sat Mar 18, 2006 6:38 pm
by Uwe Neue
I assume that you have developed methods already for content uniformity, impurity profile and degradation products of the parent drug. If this is the case, the simplest method is to develop a separate method for the counterions, maybe with ion-pair chromatography or ion chromatography, and then use this method as a generic method for the counterions.
Is this acceptable?
Posted: Sat Mar 18, 2006 7:02 pm
by zompa
Ok,
What Uwe suggested is what we do ordinarily when we analise such drugs.
We have separate methods for the counter-ions and for the "base".
So we "off-integrate" the peak of the anion while testing the counterpart purity.
bye.