We are facing some challenges in redeveloping an original HPLC method to a new UPLC method for a intermediate drug product.
Some details below.
API: C26H20FN5O2S2, MM: 517.6 g/mol (2 benzene rings, pyridine, etc...)
Intermediate drug product: API + 1 single polymer
UPLC method
Mobile Phase A: 10mM K2HPO4 Buffer, pH 7
Mobile Phase B: MeOH : ACN (50:50)
Column: BEH Phenyl 100 mm x 2.1 mm x 1.7 µm
Column temp: 30 ºC
Autosampler temp: 5 ºC
Injection volume: 5 µL
Time (min) Flow (mL/min) %Mobile Phase A %Mobile Phase B
0.00 0.3 90 10
2.00 0.3 60 40
20.00 0.3 45 55
30.00 0.3 45 55
50.00 0.3 25 75
55.00 0.3 25 75
55.10 0.3 90 10
60.00 0.3 90 10
Diluent: [DCM/EtOH 80:20, 10% + H2O/MeOH 70:30, 90%] !!!
Nominal concentration: 0.1 mg/mL API
60 min in UPLC, I know it's to long but we really need to separate the impurity before main peak. And until now....this was the "best".
Some typical and consistent profiles are presented below (Blank and Reference Standard)
UPLC profile (full scale)
[img][IMG]http://i347.photobucket.com/albums/p475/pserodio/UPLC_profile_full%20scale_zpsru4bigf3.png[/img][/img]
UPLC profile (zoomed)
[img][IMG]http://i347.photobucket.com/albums/p475/pserodio/UPLC_profile_zoomed_zpsgoh2yhhq.png[/img][/img]
UPLC pressure plot
[img][IMG]http://i347.photobucket.com/albums/p475/pserodio/UPLC_pressure%20plot_zps52vebob9.png[/img][/img]
This UPLC method provides good separation/resolution for the pair (impurity imeediately before main peak/main peak). This was not achieved with the original HPLC although this was well balanced in terms of MPA (0.1% FA in H2O) and MPB (0.1% FA in MeOH/ACN 1:1). Diluent was similar Diluent: [DCM/EtOH 80:20, 10% + 0.1% in H2O/MeOH 70:30, 90%].
Issues in UPLC:
1) To optimize the overall profile is it critical to use 10 mM phosphate buffer in MPB? To avoid dilution of buffer concentration along gradient?
2) Highest limit of 25:75 MPA:MPB ensure no buffer precipitation? Pressure plot seems "strange". Small but cyclic vertical oscillations (buffer precipitation?)
3) Unexpected(?) low absorbance at 0.1 mg/mL. Higher injection volumes tend to broaden the main peak.
4) Blank is not "well clean". There are some interfering peaks. Our COP allows only up to 10% of LOQ!!!! Starting the gradient with less MPA can eventually help but will impact on separation power.
5) Baseline depletion tend to systematically occur after the main peak. How to eliminate this?
Should we try different column? Different buffer? Any other suggestion?
Can you please take a look on this topic in detail and give back some thoughts?
Many thanks,
Pedro