Sample loss during multiple injections
Posted: Fri Mar 17, 2006 3:39 pm
I have a problem, when I inject twice out of the same vial the 2nd injection gives a signal approx 1/20th of the first. I am running HPLC-RAD and uV and the problem appears with both.
Checks I have done: contents of vial has been checked and no separation of the analyte / mobile phase appears to be happening in the vial. The vial has also been radio checked and the concentration of what remains in the vial is normal.
Following the low 2nd injection run, I have run at 95% acetonitrile for half an hour and nothing comes off the column.
I have run a different hot compound and do not have this problem, multiple injections and no significant change in signal).
What could be happening, I have had the problem the other way round (low first then higher) but not seen this before.