I'm running a free phenol % concentration on a 7890B GC. My ISTD value is showing up just fine and all peaks are appearing as normal, but my phenol % is being calculated at twice what it should be. No multiplier or dilution factor is being used.

We have run a 10% STD several times and it comes out fine, but our control and our samples are showing a phenol % higher than what we know is in them.

What could be the cause of this issue?

Thanks in advance!
Take a look at the actual chromatograms. Is there any difference in the peak shape? or the baseline?

My guess would be that you're seeing either a matrix effect distorting the peak shape and fouling up the integration (although it's hard to imagine doubling the area that way) or there are unresolved compounds in your matrix (controls & samples) coeluting with the phenol.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
Are you talking about the AREA changing or the calculation and result not giving the result that you expect. Are you asking for a ISTD report or a ISTD% report.

Gasman
If you take the peak areas and do the calculation by hand, do you get the right answer ?

If you do the problem is in the data processing, if not it is in the chromatography, sample prep or integration.

Peter
Peter Apps
4 posts Page 1 of 1

Who is online

In total there are 28 users online :: 0 registered, 0 hidden and 28 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 28 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry