Baseline

Posted: Wed Mar 15, 2006 6:33 pm
by Tammy Nguyen
Hi all,

My baseline is not a straight line. After the peak eluted, the baseline is not coming back at zero; instead, it keeps moving downward a little bit more. The peak looks like having a long tailing, and the baseline looks like having a valley. I have 1mL of triethylamine/L.

Please help me how to get rid of it.

Thank you,
TN

Posted: Thu Mar 16, 2006 12:01 am
by ntruong
Does the baseline drop in the blank run as well?

Posted: Thu Mar 16, 2006 12:42 pm
by malaka50
clean the column, maybe back flush at a low flow rate.

run gradient 95% water to 95% acetonitrile.

condition column for at least an hour, maybe longer with the TEA mobile phase.

Posted: Thu Mar 16, 2006 2:09 pm
by Tammy Nguyen
The valley drops in resolution solution only, not blank nor STD. SA. Maybe one of the components in reso. causes it.

I didn't clean the column yet since I've seen a nice blank after I ran the resolution; will try some time today.

Thanks, all.
TN

Re: Baseline

Posted: Wed Mar 22, 2006 12:32 pm
by michaldousa
I have too little information. How is composition of mobile phase (first of all pH of mobile phase)? Ion-pair reagents are use? Try to equilibrate column for a longer time.