Glycerolmonostearate by HPLC?
Posted: Fri Sep 30, 2016 8:43 am
by Mattias
I have been asked to develop an HPLC method for glycerolmonostearate (GMS) with UV and/or ESI-MS detection.
I have a validated LC/UV method for stearic acid... Not sure if it would be feasible to hydrolyse GMS to stearic acid and then use the existing method?
Or analyse GMS as it is? I suspect that GMS will have a very poor MS-response. We do not have any CAD or ELSD detectors.
Re: Glycerolmonostearate by HPLC?
Posted: Fri Sep 30, 2016 3:12 pm
by Consumer Products Guy
I've used high temperature GC after trimethylsilyl derivatization for GMS.
I have been asked to develop an HPLC method for glycerolmonostearate (GMS) with UV and/or ESI-MS detection. We do not have any CAD or ELSD detectors.
Time for you to ask for ELSD. If high enough levels of GMS, might be able to do refractive index. I would not go the saponification route myself.
Posted: Fri Sep 30, 2016 4:28 pm
by AaronAIT
LC with an RI Detector as utilised in some of the pharmacopoieal tests for GMS works quite well, as Consumer Products Guy mentioned.
Posted: Mon Oct 03, 2016 10:31 am
by Mattias
Thanks!
We have no GC and no RI or ELSD.... Just UV and MS.
It seems as the ester bond has a similar UV-absorbance as the COOH group. And I have no problem to detect stearic acid with UV (200 nm). I will try a UV-transparent mobile phase and see if I can detect GMS with UV.
Agree that the saponification route is looking like trouble...
Posted: Thu Oct 20, 2016 11:36 am
by Mattias
I tested it on my LC/UV method for stearic acid, and GMS elutes before stearic acid in that method. Seems to work fine for quantitation of GMS although the QL is about 40 µg/ml (not for trace analysis). I guess it is better than RI?
Column: XBridge C18, 3x150mm
Mobile phase: 50% ACN, 25% THF, 25% water (isocratic)
Wavelength: 200 nm
Posted: Fri Oct 21, 2016 7:56 am
by Mattias