Ion 69 on MSD 5977E

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

7 posts Page 1 of 1
Hello!

I'm a new member !! I'm working in France.

I'm using a GC/MS apparatus (7890b+MSD5977E).

We tried to change the carrier gas Helium to Hydrogen but we are getting an abundance of the 69 ion in all traces that we cannot tune out, this affects the library identification.
The standard spectra tune seems to be ok.(see below the pic)

May you help me please ?

Best regards

EmmanuelImage
Nobody got an idea please :roll:
If I understand you correctly you have ion 69 in all your samples in the backgroung.

Maybe PFTBA valve is leaking ?
That's right in each samples...
Gonna check this valve

Thanks
If you switch from Helium to Hydrogen you will have quite a large amount of background ions when you first begin doing analysis. It takes a few weeks for the Hydrogen to clean out any and all contaminates from the gas flow path(it cleans really well) then once it reaches an equilibrium you will see better background.

If you scan in manual tune with the calibration gas valve closed, you will probably see a lot of ions present at first, over time they will be less and less. It looks like hydrocarbon background is high until it flushes out.
The past is there to guide us into the future, not to dwell in.
If you switch from Helium to Hydrogen you will have quite a large amount of background ions when you first begin doing analysis. It takes a few weeks for the Hydrogen to clean out any and all contaminates from the gas flow path(it cleans really well) then once it reaches an equilibrium you will see better background.
Can you use this to clean your flow path, even if you have no plans to switch to hydrogen for the analysis?
I ran one instrument on Hydrogen for a month or so, and when I switched back to Helium it had a super clean background when scanning with the cal valve closed. It would seem to work quite well to clean up a system, just takes a long time to do it. If you were to run Hydrogen any time the instrument was in standby, that would probably help keep things cleaner than normal I would imagine.
The past is there to guide us into the future, not to dwell in.
7 posts Page 1 of 1

Who is online

In total there are 15 users online :: 1 registered, 0 hidden and 14 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Amazon [Bot] and 14 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry