Method validation for GPC method
Posted: Tue Sep 27, 2016 3:00 pm
by michaelcarolus
Hi All
I hope you can help me. I have a GPC method that I need to validate. The method allows for the quantitative determination of the average molecular weight and molecular weight distribution in my samples.
I was thinking of doing the following for my method validation:
    Robustness - variation in mobile phase pH, buffer concentration of mobile phase, variation in column temperature, variation in inj vol, variation in flow rate, variation in UV wavelength.
      Method Precision - Repeatability and Intermediate Precision
      Are there any other "typical" validation parameters that i should consider? Are they any other available guidelines that can help me to better understand the requirements for validating a GPC method?
      Thanks in advance
      Mike
      Re: Method validation for GPC method
      Posted: Thu Sep 29, 2016 1:49 pm
      by danko
      Accuracy!
      Use a known standard or just some substance you have a documented Mw and see if you can determine the same Mw using your method.