Sorry for my bad English)
I have a trouble with Agilent 5973 MSD. It started when power outage happened and MSD powered off during analysis.
After powered on, Autotune stopped with error "Tune action stopped: Gain and offset = 0 and peaks are too narrow!"
I plugged new filament. No changes.
I cleaned the ion source. No changes.
I tested voltage and resistance of filaments. It was normal.
And then I pinged nests of filaments, repeller, ion focus and entrance on green board (J2 and J5 places on photo) with ground with my multimeter.

And all are pinged except ion focus.
I did it same with working Agilent 5975С. There are no ping on ion focus and entrance!!!
I decided, that trouble with entrance in MSD 5973.