Ion 69 on MSD 5977E
Posted: Fri Sep 23, 2016 7:28 am
by Emmanuel
Hello!
I'm a new member !! I'm working in France.
I'm using a GC/MS apparatus (7890b+MSD5977E).
We tried to change the carrier gas Helium to Hydrogen but we are getting an abundance of the 69 ion in all traces that we cannot tune out, this affects the library identification.
The standard spectra tune seems to be ok.(see below the pic)
May you help me please ?
Best regards
Emmanuel
Re: Ion 69 on MSD 5977E
Posted: Mon Sep 26, 2016 9:08 am
by Emmanuel
Nobody got an idea please
Posted: Mon Sep 26, 2016 12:51 pm
by dblux_
If I understand you correctly you have ion 69 in all your samples in the backgroung.
Maybe PFTBA valve is leaking ?
Posted: Mon Sep 26, 2016 2:09 pm
by Emmanuel
That's right in each samples...
Gonna check this valve
Posted: Tue Sep 27, 2016 9:39 pm
by James_Ball
If you switch from Helium to Hydrogen you will have quite a large amount of background ions when you first begin doing analysis. It takes a few weeks for the Hydrogen to clean out any and all contaminates from the gas flow path(it cleans really well) then once it reaches an equilibrium you will see better background.
If you scan in manual tune with the calibration gas valve closed, you will probably see a lot of ions present at first, over time they will be less and less. It looks like hydrocarbon background is high until it flushes out.
Posted: Wed Sep 28, 2016 6:54 am
by Rndirk
Can you use this to clean your flow path, even if you have no plans to switch to hydrogen for the analysis?
Posted: Thu Sep 29, 2016 7:27 pm
by James_Ball
I ran one instrument on Hydrogen for a month or so, and when I switched back to Helium it had a super clean background when scanning with the cal valve closed. It would seem to work quite well to clean up a system, just takes a long time to do it. If you were to run Hydrogen any time the instrument was in standby, that would probably help keep things cleaner than normal I would imagine.