Single Point calibration Hydrocarbon Analysis

Discussions about GC and other "gas phase" separation techniques.

2 posts Page 1 of 1
We are setting up two methods of analysis for hydrocarbons (gas and liquid methods) for the first time on new instruments in a new lab. I believe a single point is insufficient and inaccurate for determination. Single point calibration is only suitable when concentration is very close to the actual sample therefore there is less potential for error. I believe we have a couple of options
1. Low-Medium-High concentration's of standard which will give us an accurate cali curve
2. three points with a concentration very close to the expected concentration of the sample.
Basically im looking for the most accurate way of quantifying with the least possible error

Just wondering what are your thoughts on this?
Single point with concentration close to sample can work well if you use internal standard with samples and standards and calculating concentration by area ratio. To proof method precision, LOD/LOQ 6 replicates of the standard can be run, area RSD should be within 5%.
For polar carbohydrates we use LC/MS.
2 posts Page 1 of 1

Who is online

In total there are 29 users online :: 1 registered, 0 hidden and 28 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Bing [Bot] and 28 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry