Hi
This is my first post to this forum so please excuse me if miss someting.

We are setting up a GC system to check residual solvent, however we are having some headaches.
We have made a standard solution consisting of 0,1% ethanol and 0,01% acetonitrile. Analysing this solution with our system, we are finding that the peaks from ethanol are behaving much more erratic than acetonitrile. In Our last run we had six samples, yielding a 16% RSD for ethanol and 7% for acetonitrile. This is a very typical result. We have recently had service on the system, and both injector and FID has been replaced. Switching to a new column had no effect. What are we missing. Why are ethanol giving so poor precision of the peak-area of ethanol?

The system is GC-FID
1:3 split
140 oC injector, 250 oC detektor.
program;
5 min @ 50 oC
10 degrees/min to 130 oC
2,7 ml/min helium flow

Head Space PAL system
10 ml head Space vial
4950 µl water 50 µl sample
45 min @ 80 oC incubation
90 oC Needle temp
1 ml injection volume

Syring flushed With helium between injections
It's been a while for my doing HS blood alcohol, but as I remember I started my oven at 35 or 40. I would bump your injector to 200 or so.
We injected 100 ul of sample without splitting.
You are probably getting a pressure pulse as you inject - a 1ml sample volume is 25% of the total flow and would need an injection time of several seconds to avoid a pulse. Either reduce the injection volume or increase the split ratio and see what happens.

Peter
Peter Apps
The injection speed is 1000 µl/sec. Could this be increased to say, 330µl/sec? I've tried to up the split to 1:5 without much effect. RSD for ethanol is very poor. We used to have much higher flow, i.e. 10 ml/min, however no other effect than lower retention times.
The injection speed is 1000 µl/sec. Could this be increased to say, 330µl/sec? I've tried to up the split to 1:5 without much effect. RSD for ethanol is very poor. We used to have much higher flow, i.e. 10 ml/min, however no other effect than lower retention times.
1000 ul/sec is 60 ml/min so you are definitely generating a pressure pulse that causes variability in transfer of sample to the column.

Increase the split to 20:1 or decrease the injection volume to 100 ul.

Peter
Peter Apps
Thanks for the input
Increasing the injector to 200 oC cant be done, as we are using the PhEur method B for our system. Playing around with the split is OK.
I adjusted the split to 1:10 (an also reduced injection speed to 300 µl/sec), however the answers are much the same.

Running nine injections of the same test solution gave
5.0% RSD for THF (mean 46,7 uV.Min),
11.1% RSD for EtOH (mean 266.1 uV.Min)
5.3% RSD for AcCN (mean 38.3 uV.Min)

Which all are below the PhEur requirement of 15% RSD, but still not what I would call good.
Which leads me to the quiestion: What RSD can be demanded of such a system? Is RSD below 5% for repeated injections to much to ask for, for a PAL-CTC GC-FID system?
7 posts Page 1 of 1

Who is online

In total there are 41 users online :: 0 registered, 0 hidden and 41 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 41 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry