I'm using a shimadzu GC-14
restek rt Q Plot 30 m column.
detector 220
inj 200
initial temp 45
initial time 2
prog rate 10
final temp 160
final time 6.5
total time 20
I have two different type of standards that I purchased. The first is a gas cylinder that contains 1000 ppm each of butane, ethane, hexane, methane,pentane, & propane.
The second standard contains 100 ppm of acetone, acetonitrile, etoh, 2 propanol, meoh, propane, butane, isobutane, hexane, and heptane in DMSO.
The samples that I will be testing will be herbal extracts. I will weigh out ~50mg into a 40ml voa, heat the voa in a heating block, and using a gas tight syringe I will pull up 500ul of headspace and manually inject into the gc.
For my curve if I use the gas cylinder standard I have been with drawing 500ul directly into a syringe and injecting that directly into the gc.
Using peak simple I set the area of each peak to equal 1000ppm
To calculate the result of a sample that has a butane peak I am doing the following calculation

Does this seem correct?
My other question is if I use the standard that is dissolved in dmso instead of the gas cylinder, would I inject it in the liquid state, or would I treat it like a sample and put it in a headspace voa, heat it, and then inject it. If I were to put it into a voa, how much would I put into the voa? I believe that I would want the injection amount to stay the same as with a sample, but im not sure what the ppm of standard would be once placed in the voa.
Am I making any sense at all?