High 44 background

Posted: Wed Sep 14, 2016 6:20 pm
by Yama001
I have another 5972 that has always plagued the operator with high 44 background, that will recede after several days of pumping down. I am fairly convinced this is not a leak (carbon dioxide), as we do not see associated 28, 32, or 40 ions. We do not see this on other instruments on the same gas line, so I doubt it is the tank.

This is a diffusion pump system. I do not believed 44 is associated with the pump fluid? I know it has not been changed in years, so I am trying to talk myself into it.

Any ideas?

Re: High 44 background

Posted: Wed Sep 14, 2016 8:41 pm
by James_Ball
I have another 5972 that has always plagued the operator with high 44 background, that will recede after several days of pumping down. I am fairly convinced this is not a leak (carbon dioxide), as we do not see associated 28, 32, or 40 ions. We do not see this on other instruments on the same gas line, so I doubt it is the tank.

This is a diffusion pump system. I do not believed 44 is associated with the pump fluid? I know it has not been changed in years, so I am trying to talk myself into it.

Any ideas?
Could be the old oil has some carbon built up in it and every time you cool and vent it absorbs some CO2, then bakes out after a few days. Good time to change the oil, and do a good cleaning on the components of the diff stack while you are in there.

Re: High 44 background

Posted: Thu Sep 15, 2016 2:11 pm
by Yama001
Thanks James. I may replace the pump fluid next time I can afford to take that instrument down.

Re: High 44 background

Posted: Mon Sep 26, 2016 4:24 am
by mckrause
This is a common problem with graphite ferrules. Never completely understood why, but we would see 44 (CO2) if we overtightened the graphite ferrule at the MS interface. Switched to aluminum ferrules and all of the problem went away.

Re: High 44 background

Posted: Mon Sep 26, 2016 6:52 am
by Rndirk
This is a common problem with graphite ferrules. Never completely understood why, but we would see 44 (CO2) if we overtightened the graphite ferrule at the MS interface. Switched to aluminum ferrules and all of the problem went away.
Interesting. Are you referring to these ferrules (Link)? I haven't tried them. Do they perform as promised, e.g no re-tightening, more leak-free?

Re: High 44 background

Posted: Tue Sep 27, 2016 2:25 pm
by BMU_VMW
Do they perform as promised, e.g no re-tightening, more leak-free?
Yes they do. We use them to connect retention-gap and analytical column and that works perfect.
Have not yet tried to use them for a transferline.